Posted by Edward d'Auvergne on July 20, 2009 - 14:18:
Hi,

Sorry for the delayed response, the last few months for me have been
crazy.  Please see below for more:


On Tue, Jun 2, 2009 at 8:36 PM, Sébastien
Morin<sebastien.morin.1@xxxxxxxxx> wrote:
Hi,

This is a fairly old post, but I finally had time to think about these
issues... Please see below...



Edward d'Auvergne wrote:

Hi,

Is the frequency for the reference spectrum necessary?  Isn't
cmpg_delayT set to zero in this case, i.e. the CPMG block is missing?
If it is necessary though, a value of None is probably a better choice
to identify it rather than the frequency of zero Hz.

I guess recording a reference for each frequency is necessary since the
intensity is to be extracted and could vary when changing magnet (along with
S/N)...

I agree for a value of None for the reference spectrum (which is what is
presently in the code).

Ok, so we need a reference frequency set, but cmpg_delayT set to None.


Another question I have is should the nu_cmpg value be given (with Hz
units), or would it be better if the omega_cmpg value was given (with
rad/s units)?  If nu_cmpg is given, this will have to be converted
later to omega.  I think we should have an explanation of both, after
the relevant model equations.  Also the 'frq' arg of
relax_disp.cpmg_frq() might be better named as nu_cmpg or omega_cmpg
for clarity if this is frequency or angular frequency.

For this part, I am not sure about the units to use... 'cpmg_frq' needs to
be of the same units as 'kex' and 'dw' (see equations below). I guess 'kex'
and 'dw' should be in rad/s, so 'cpmg_frq' should also be in rad/s...

Is it right ?

Depending on the answer, 'cpmg_frq' will be renamed (to either 'cpmg_nu' or
'cpmg_omega').

I think we should use omega units (with the hidden radian unit).  Do
you know what is normally used?



--------------------------

FAST EXCHANGE

                  /              /        kex       \    4 * cpmg_frq \
R2eff = R2 + Rex * | 1 - 2 * Tanh | ------------------ | * ------------- |
                  \              \ 2 * 4 * cpmg_frq /         kex     /

SLOW EXCHANGE
                  /     /      dw      \    4 * cpmg_frq \
R2eff = R2A + kA - | Sin | -------------- | * ------------- |
                  \     \ 4 * cpmg_frq /         dw      /

where cpmg_frq = 1 / ( 4 * cpmg_tau ).


Also note that
we have to convert the cmpg_delayT value too.  Unit analysis of the
equation

R2eff = ( 1 / T ) * Ln( Icpmg / Iref )

shows this.  R2 is in units of rad/s.  T is input in seconds.  1/T is
frequency in nu units of Hz.  Therefore we need to convert to the
radian units of angular frequency by multiplying by 2pi as 2pi/T is in
rad/s units.  The natural logarithm of peak intensities is unitless
and dimensionless.

I just had a look at the reference dataset included in the test suite (from
Hansen et al., J. Phys. Chem., 2008)...

When treating the delay T as is (in seconds), I get the same values for
R2eff as published in the paper (for the FF domain). However, if multiplying
the delay T by 2pi, I get values for R2eff that a way too big.

delay T is in the pulse sequence and should be in seconds.


I do not want to say that the logic behind unit analysis is deficient. I
agree with that logic, but I also think that, in this case, the delay T
should stay in seconds in order to get R2eff values of the good size...

Despite delay T being in seconds, R2eff is in rad/s.  This is the same
as standard R1 or R2 where the time period in the pulse sequence is in
seconds whereas the fitted rate is in rad/s.


What do you think ?

As long as a number of published results can be replicated, we should be fine.