Posted by Chris MacRaild on October 17, 2008 - 01:59:
On Thu, Oct 16, 2008 at 8:07 PM, Edward d'Auvergne
<edward.dauvergne@xxxxxxxxx> wrote:
On Thu, Oct 16, 2008 at 7:02 AM, Chris MacRaild <macraild@xxxxxxxxxxx> 
wrote:
On Thu, Oct 16, 2008 at 3:11 PM, Sébastien Morin
<sebastien.morin.1@xxxxxxxxx> wrote:
Hi,

I have a general question about curve fitting within relax.

Let's say I proceed to curve fitting for some relaxation rates
(exponential decay) and that I have a duplicate delay for error 
estimation.

========
delays

0.01
0.01
0.02
0.04
...
========

Will the mean value (for delay 0.01) be used for curve fitting and rate
extraction ?
Or will both values at delay 0.01 be used during curve fitting, hence
giving more weight on delay 0.01 ?

In other words, will the fit only use both values at delay 0.01 for
error estimation or also for rate extraction, giving more weight for
this duplicate point ?

How is this handled in relax ?

Instinctively, I would guess that the man value must be used for
fitting, as we don't want the points that are not in duplicate to count
less in the fitting procedure... Am I right ?


I would argue not. If we have gone to the trouble of measuring
something twice (or, equivalently, measuring it with greater
precision) then we should weight it more strongly to reflect that.

So we should include both duplicate points in our fit, or we should
just use the mean value, but weight it to reflect the greater
certainty we have in its value.

As I type this I realise this is likely the source of the sqrt(2)
factor Tyler and Edward have been debating on a parallel thread - the
uncertainty in height of any one peak is equal to the RMS noise, but
the std error of the mean of duplicates is less by a factor of
sqrt(2).

At the moment, relax simply uses the mean value in the fit.  Despite
the higher quality of the duplicated data, all points are given the
same weight.  This is only because of the low data quantity.  As for
dividing the sd of differences between duplicate spectra by sqrt(2),
this is not done in relax anymore.  Because some people have collected
triplicate spectra, although rare, relax calculates the error from
replicated spectra differently.  I'm prepared to be told that this
technique is incorrect though.  The procedure relax uses is to apply
the formula:

sd^2 = sum({Ii - Iav}^2) / (n - 1),

where n is the number of spectra, Ii is the intensity in spectrum i,
Iav is the average intensity, sd is the standard deviation, and sd^2
is the variance.  This is for a single spin.  The sample number is so
low that this value is completely meaningless.  Therefore the variance
is averaged across all spins (well due to a current bug the standard
deviation is averaged instead).  Then another averaging takes place if
not all spectra are duplicated.  The variances across all duplicated
spectra are averaged to give a single error value for all spins across
all spectra (again the sd averaging bug affects this).  The reason for
using this approach is that you are not limited to duplicate spectra.
It also means that the factor of sqrt(2) is not applicable.  If only
single spectra are collected, then relax's current behaviour of not
using sqrt(2) seems reasonable.


Here is how I understand the sqrt(2) issue:

The sd of duplicate (or triplicate, or quadruplicate, or ... ) peak
heights is assumed to give a good estimate of the precision with which
we can measure the height of a single peak. So for peak heights that
have not been measured in duplicate (ie relaxation times that have not
been duplicated in our current set of spectra), sd is a good estimate
of the uncertainty associated with that height.

For peaks we have measured more than once, we can calculate a mean
peak height. The precision with which we know that mean value is given
by the std error of the mean ie. sd/sqrt(n) where n is the number of
times we have measured that specific relaxation time. I think this is
the origin of the sqrt(2) for duplicate data.

A made up example:
T           I
0          1.00
10        0.90
10        0.86
20        0.80
40        0.75
70        0.72
70        0.68
100      0.55
150      0.40
200      0.30

The std deviation of our duplicates is 0.04 so the uncertainty on each
value above is 0.04

BUT, the uncertainty on the mean values for our duplicate time points
(10 and 70) is 0.04/sqrt(2) = 0.028

So if we use the mean values as points in our fit, we should use 0.028
as the uncertainty on those values (while all other peaks have
uncertainty 0.04)

Alternatively (and equivalently) we can use the original observations,
including all duplicate points. In this case, all points have the same
uncertainty of 0.04, as they are all the result of a single
measurement.

To do anything else is to underestimate the precision with which we
have measured our relaxation rates.

Chris




Regards,

Edward


P.S.  The idea for the 1.3 line is to create a new class of user